Capacitance-voltage (C-V) testing is widely used to determine
semiconductor parameters, particularly in MOSCAP and MOSFET
structures. However, other types of semiconductor devices and
technologies also can be characterized with C-V measurements,
including bipolar junction transistors, JFETs, III-V compound
devices, photovoltaic cells, MEMS devices, organic thin film
transistor (TFT) displays, photodiodes, and carbon nanotubes.
The fundamental nature of these measurements makes them relevant to
a wide range of applications and disciplines. They are used in the
research labs of universities and semiconductor manufacturers to
evaluate new materials, processes, devices, and circuits. C-V
measurements are extremely important to product and yield
enhancement engineers responsible for improving processes and
device performance. Reliability engineers use these measurements to
qualify material suppliers, monitor process parameters, and analyze
failure mechanisms.
With appropriate methodologies, instrumentation, and software, a
multitude of semiconductor device and material parameters can be
derived. This information is used all along the production chain
beginning with evaluation of epitaxially grown crystals including
parameters such as average doping concentration, doping profiles,
and carrier lifetimes.
In wafer processes, C-V measurements can reveal oxide thickness,
oxide charges, contamination from mobile ions, and interface trap
density. These measurements continue to be important after other
process steps such as lithography, etching, cleaning, dielectric
and polysilicon depositions, and metallization. After devices are
fully fabricated on the wafer, C-V is used to characterize
threshold voltages and other parameters during reliability and
basic device testing and to model the performance of these devices.
The Physics of Semiconductor Capacitance
A MOSCAP structure is a fundamental device formed during
semiconductor fabrication (Figure 1). Although these devices may be used in actual circuits, they
typically are integrated into fabrication processes as test
structures. Since they are simple structures and fabrication is
easy to control, they are a convenient way to evaluate the
underlying processes.

Figure 1. C-V Measurement Circuit for a MOSCAP Structure Formed on
a P-Type Substrate
The metal/polysilicon layer shown in Figure 1 is one plate of the
capacitor, and silicon dioxide is the insulator. Since the
substrate below the insulating layer is a semiconducting material,
it is not by itself the other plate of the capacitor. In effect,
the majority charge carriers become the other plate.
Physically, capacitance (C) is determined from the variables in the
following equation:
C = A (k/d)
where: A = area of the capacitor
k = dielectric constant of the insulator
d = separation of the two plates
As a result, the larger A and k are and the thinner the insulator
is, the higher the capacitance will be. Typically, semiconductor
capacitance values range from nanofarads to picofarads or smaller.
The procedure for taking C-V measurements involves the application
of DC bias voltages across the capacitor while making the
measurements with an AC signal (Figure 1). Commonly, AC frequencies
from about 10 kHz to 10 MHz are used for these measurements. The
bias is applied as a DC voltage sweep that drives the MOSCAP
structure from its accumulation region into the depletion region
and then into inversion (Figure 2).

Figure 2. DC Bias Sweep of MOSCAP Structure Obtained During C-V
Testing
A strong DC bias causes majority carriers in the substrate to
accumulate near the insulator interface. Since they can’t get
through the insulating layer, capacitance is at a maximum in the
accumulation region as the charges stack up near that interface
because d is at a minimum (Figure 1). One of the fundamental
parameters that can be derived from C-V accumulation measurements
is the silicon dioxide thickness.
As the bias voltage decreases, majority carriers get pushed away
from the oxide interface, and the depletion region forms. When the
bias voltage is reversed, charge carriers move the greatest
distance from the oxide layer, and capacitance is at a minimum
because d is at a maximum. From this inversion region capacitance,
the number of majority carriers can be derived. The same basic
concepts apply to MOSFETs even though their physical structure and
doping are more complex.
Many other parameters can be derived from the three regions shown
in Figure 2 as the bias voltage is swept through them. Different AC
signal frequencies can reveal additional details. Low frequencies
uncover what are called quasistatic characteristics where
high-frequency testing is more indicative of dynamic performance.
Both types of C-V testing often are required.
Basic Test Setup
Figure 3 is the block diagram of a basic C-V measurement setup. Because C-V
measurements actually are made at AC frequencies, the capacitance
for the DUT is calculated with the following:

where: IDUT = magnitude of the AC current through the DUT
f = test frequency
VAC = magnitude and phase angle of the measured AC voltage

Figure 3. Basic Test Setup for C-V Measurements
In other words, the test measures the AC impedance of the DUT by
applying an AC voltage and measuring the resulting AC current, AC
voltage, and impedance phase angle between them.
These measurements take into account series and parallel resistance
associated with the capacitance as well as the dissipation factor. Figure 4 illustrates the basic circuit variables that can be derived from
the measurements.

Figure 4. Basic Electrical Variables Available From C-V
Measurements
Challenges of C-V Measurements
Although a block diagram of a C-V test setup looks deceptively
simple, certain challenges are associated with this testing.
Typically, test personnel have problems in the following areas:
• Low-capacitance measurements of picofarads and smaller values
• C-V instrument connections through a prober to the wafer device
• Leaky, high D capacitance measurements
• Using hardware and software to acquire the data
• Parameter extractions
Overcoming these challenges requires careful attention to the
techniques used along with appropriate hardware and software.
Low-Capacitance Measurements
If C is small, the DUT’s AC response current is small and hard to
measure. However, at higher frequencies, the DUT impedance is
reduced so the current increases and is easier to measure.
Often, semiconductor capacitance is less than 1 pF, which is below
the capabilities of many LCR meters. Even those claiming to measure
these small capacitance values may have confusing specifications
that make it difficult to determine the final accuracy in the
measurement. If accuracy over the instrument’s full measurement
range is not explicitly stated, you need to clarify this with the
manufacturer.
C-V Measurement Connections
In most test environments, the DUT is a test structure on a wafer:
It is connected to the C-V instrument through a prober, a probe
card adapter, and a switch matrix. Even if no switch is involved,
there still is a prober and significant cabling.
At high frequencies, special corrections and compensation must be
applied. Usually, this is achieved with some combination of an
open, short, or calibration device.
Because of the complexity of the hardware, cabling, and
compensation techniques, it is a good idea to confer with C-V test
application engineers. They are skilled at working with various
probe systems to overcome many types of interconnection problems.
High D (Leaky) Capacitors
In addition to having a low C value, a semiconductor capacitor also
may be leaky. That is the case when the equivalent R in parallel
with C is too low. This results in resistive impedance overwhelming
the capacitive impedance, and the C value gets lost in the noise.
For devices with ultrathin oxide layers, D values can be greater
than five. In general, as D increases, the accuracy of a C
measurement is rapidly degraded so high D is a limiting factor in
the practical use of a C meter. Again, higher frequencies can help
solve the problem. At higher frequencies, the capacitive impedance
is lower, resulting in a C current that is higher and more easily
measured.
Obtaining Useful Data
In addition to accuracy issues, practical considerations in C-V
data collection include the instrumentation’s range of test
variables, versatility of parameter extraction software, and ease
of hardware usage. Traditionally, C-V testing has been limited to
about 30 V and 10 mA DC bias. However, many applications such as
characterizing laterally diffused MOS structures, low-k interlayer
dielectrics, MEMS devices, organic TFT displays, and photodiodes
require tests at higher voltage or current. For these applications,
a separate high-voltage DC power supply and C meter are required;
DC bias from 0 to ±400 V and a current output up to 300 mA are very
useful. Being able to apply differential DC bias on both the HI and
LO terminals of the C-V instrument offers more flexible control
over electric fields within the DUT, which is very helpful in
researching and modeling novel devices such as nanoscale
components.
The instrumentation software should include ready-to-run test
routines that do not require user programming. These should be
available for the most widely used device technologies and test
regimens.
Some researchers also may be interested in less common tests such
as performing both a C-V and C-f sweep on a metal-insulator-metal
capacitor, measuring small interconnect capacitance on a wafer, or
doing a C-V sweep on a two-terminal nanowire device. The parameter
extractions should be easily obtained with automated curve
plotting.
Often, engineers and researchers are expected to perform C-V
measurements with little experience and training on the
instrumentation. A test system with an intuitive user interface and
easy-to-use features makes this practical. Simple test setup,
sequence control, and data analysis are essential. Otherwise, you
spend more time learning the system than collecting and using the
data.
Other considerations for a test system include:
• Tightly integrated source-measure units, a digital oscilloscope,
and a C-V meter
• Easy integration with other external instruments
• DC biasing down to millivolts and capacitance measurements down
to femtofarads to support high-resolution, precise measurements at
the probe tips
• Test setups and libraries that can be easily modified
• Diagnostic/troubleshooting tools that let you know whether or not
the system is performing correctly
About the Author
Lee Stauffer is a senior marketer responsible for developing and
supporting products for the semiconductor manufacturing and
research markets at Keithley Instruments. His formal education in
electrical engineering and semiconductor device physics is
complemented by 20 years experience in semiconductor process and
product engineering, device characterization, and instrumentation
design. Keithley Instruments, 28775 Aurora Rd., Cleveland, OH
44139, 440-248-0400, e-mail: lstauffer@keithley.com